Medical Endoscope Cold Light Source Testing

Medical endoscope cold light sources are tested per YY 1081-2011 at the reference window for Ra, CCT, flux ratios and IR cutoff. SPM-8000Y with fibre and sphere covers colorimetry; IUT-500/PH-200 handles illuminance uniformity.

Definition & Scope

Medical endoscope cold light sources supply illumination for endoscopic imaging; their photometric and colorimetric performance affects clinical image quality and safety. YY 1081-2011 specifies parameters at the reference window: P(λ), chromaticity, CCT Tc, CRI Ra/Ri, peak wavelength and FWHM, luminous/radiant flux, red/green/blue flux ratios, IR cutoff φe/φ(300–1700 nm), plus illuminance uniformity and exceedance points. Spectral colorimetry typically uses SPM-8000Y with a BaSO4 integrating sphere; uniformity testing uses the IUT-500 medical cold light uniformity system centred on the PH-200 photometer with dedicated fixtures and software for uniformity, exceedance count, average illuminance and relative sample standard deviation.

FAQ

Which core parameters does YY 1081-2011 require?
At minimum: P(λ), chromaticity and CCT, CRI Ra/Ri, peak wavelength and FWHM, luminous/radiant flux, red/green/blue flux ratios φer/φeg and φeb/φeg, and IR cutoff φe/φ(300–1700 nm), all at the reference window.
Why are standard fibre and the reference window required?
YY 1081-2011 defines reference-window geometry and standard-fibre output conditions so measurements are comparable and traceable across labs and batches. SPM-8000Y includes fibre meeting the standard.
How are Ra and red/green/blue flux ratios judged?
Typical limits include Ra≥90, CCT 3000–7000 K, and specified ranges for red/green/blue flux ratios relative to green. Exact limits follow YY 1081-2011 and product specs; software should export registration-ready reports.
Is illuminance uniformity testing also needed?
YY 1081-2011 also covers output uniformity. Use SPM-8000Y with an integrating sphere for spectral colorimetry; illuminance uniformity, exceedance count, average illuminance and relative sample standard deviation are measured with the IUT-500 system using the PH-200 photometer.